What's New Unreplied Topics Membership About Us Contact Us Privacy Policy
[Ad]

Standards for Measuring Device Calibration

Started by , Mar 24 2017 06:11 AM
4 Replies

Hello everybody,

 

We are using thermometers, micrometer, vernier caliper and weighing scale in our production. All are calibrated by a third party calibration laboratory. We also conducted in-house calibration?

 

My question are: 

 

1.How do we know that the measuring devices mentioned above is already out of tolerance limit when we conduct in-house calibration and it should be reported as un-usable? 

2. Is it based on the size of error/deviation from reference standard? By how much? e.g. 5 unit higher or lower?

3. Are there any regulations or guidelines that we can based our decision to retain or to discard the measuring devices?

 

Hoping for your help.

 

regards,

redfox

 

 

 

Share this Topic
Topics you might be interested in
How do you feel about the effectiveness of unannounced audits in upholding GFSI standards? Are Shorts Permitted Under BRC Standards for Packaging Facilities? Are there any food safety standards or guidelines for Freeze Dried Fruits and Vegetables? Documenting non-applicable standards Tesco Audit Standards
[Ad]

Hi Redfox

 

The Companies which are involved in calibration of your measuring devices should issue certification of each measuring device. These certificates  should also explain the specification which the device is calibrated to including  tolerance limit

 

Kind regards

Dr Humaid Khan 

Managing Director

Halal International Services

Sydney Australia

1 Thank

Hello Dr. Humaid Khan,

 

Thank you for your reply.

 

Assuming that our measuring devices have been certificated as "good" and it has a 1 year calibration validation, but then during in-house calibration there is a deviation found, by how much degree/size of deviation that we can tell the devices is out of tolerance already and should not be used. If the size of deviation is big e.g. 6 degrees Celcius but we apply the correction factor, is it still be acceptable?

 

regards,

redfox

Hello Dr. Humaid Khan,

 

Thank you for your reply.

 

Assuming that our measuring devices have been certificated as "good" and it has a 1 year calibration validation, but then during in-house calibration there is a deviation found, by how much degree/size of deviation that we can tell the devices is out of tolerance already and should not be used. If the size of deviation is big e.g. 6 degrees Celcius but we apply the correction factor, is it still be acceptable?

 

regards,

redfox

 

Hi redfox,

 

This is Metrology which is a science unto itself. The comments below are probably simplifications.

 

The desired performance of a given measuring instrument relates to its usage, ie what you are measuring and to what required accuracy.

 

The typical "accuracy"  delivered by measuring intruments is defined by their specifications (and often by cost). Often expressed by a maximum percentage error as a function of measurement value over defined ranges. Statistics can provide more elaborate answers.

 

The capability of a measuring instrument to meet its specification can be verified/validated by Calibration. And Statistics.

 

Nonetheless, there are typical "rules-of-thumb" for various instruments/functions which are commonly stated in calibration SOPs. These may be usable if the Procedure is validated and matches yr own usage requirements

 

As per post2, a comprehensive thermometer calibration certificate allows you to estimate the accuracy of its indicated readings for the test points used and predict the accuracy within the range employed.

 

Regarding post 3, if the unit referred is a thermcouple combination, being calibrated in ice-water and with no reset option i would first repeat with another, known accurate, probe. If same result send combination to the repair shop. If second result OK, discard the first probe and send it to the repair shop.

1 Thank

Thank you charles....

 

regards,

redfox


Similar Discussion Topics
How do you feel about the effectiveness of unannounced audits in upholding GFSI standards? Are Shorts Permitted Under BRC Standards for Packaging Facilities? Are there any food safety standards or guidelines for Freeze Dried Fruits and Vegetables? Documenting non-applicable standards Tesco Audit Standards GFSI standards that include both manufactured and traded goods. Inclusion of Climate Change To Management system Standards European Standards for Food: Ensuring Safety and Quality Harmonised Standards and Food Safety: Ensuring Quality and Compliance Inclusion of Climate Change to Management System Standards