Jump to content

  • Quick Navigation
Photo

Validation of Foreign Body Detection Device

Share this

  • You cannot start a new topic
  • Please log in to reply
10 replies to this topic
- - - - -

Robert Rogers

    Grade - AIFSQN

  • Corporate Sponsor
  • 36 posts
  • 28 thanks
5
Neutral

  • United States
    United States
  • Gender:Male

Posted 06 March 2012 - 06:38 PM

What procedure did you utilize to validate your Foreign Body Detection device?



Charles.C

    Grade - FIFSQN

  • IFSQN Moderator
  • 20,542 posts
  • 5673 thanks
1,548
Excellent

  • Earth
    Earth
  • Gender:Male
  • Interests:SF
    TV
    Movies

Posted 08 March 2012 - 12:31 PM

What procedure did you utilize to validate your Foreign Body Detection device?


Dear Robert Rogers,

What kind of device ?
Product / Process ?
Any particular standard ? It can matter.

Rgds / Charles.C

Kind Regards,

 

Charles.C


Robert Rogers

    Grade - AIFSQN

  • Corporate Sponsor
  • 36 posts
  • 28 thanks
5
Neutral

  • United States
    United States
  • Gender:Male

Posted 08 March 2012 - 01:04 PM

Metal Detectors or X-ray systems

during the recent GFSI event spoke to several auditors. It was brought to my attention that a frequent non-conformance issued is regarding the lack of validation for their foreign body detection device.

Therefore, I wanted to see what others on the site have done to satisfy this requirement



Have they used any SOP such as:



Ensure equipment is installed properly

System capable of operating to maximum sensitivity with no affect from the surrounding environment or equipment

Establish production sensitivity so that there are no false rejects

Determine achievable detection capability by passing control samples through the weakest area for detection



Charles.C

    Grade - FIFSQN

  • IFSQN Moderator
  • 20,542 posts
  • 5673 thanks
1,548
Excellent

  • Earth
    Earth
  • Gender:Male
  • Interests:SF
    TV
    Movies

Posted 08 March 2012 - 03:58 PM

Dear Robert Rogers,

You have mentioned 2 functions, sop, validation which tend to be (haccp) partitioned. In reality they can obviously overlap.

I assume the auditors/NCs you mention were referring to a haccp implementation.

Can personally only refer to metal detectors (MD). IMEX, haccp-wise, the validation aspect tends to be (narrowly) associated with validation of critical limits typically for a MD/CCP via, for example, use of (“wand”) standards. This function should clearly involve things like sensitivity maintenance aspects, but practically not always the case IMEX, ie (+) or (-) test(s). Or perhaps sensitivity etc checks are outsourced as a (annual?) certification “sop”.

There are also a range of viewpoints here regarding the exact status of the detector, eg as a CCP itself, a monitor, a verification tool . May depend on the process.

A popular operational reference for Metal detectors is the Codex validation procedure. Just search (or google) "codex validation". One recurring difficulty if you compare standards and forum threads is as to the variety of interpretations of validation.

Rgds / Charles.C


Kind Regards,

 

Charles.C


Robert Rogers

    Grade - AIFSQN

  • Corporate Sponsor
  • 36 posts
  • 28 thanks
5
Neutral

  • United States
    United States
  • Gender:Male

Posted 09 March 2012 - 05:54 PM

Thanks Charles C.

Codex mentions three areas Validation, Verification and Monitoring.

It seems

Validation would be the process to determine what level of control is possible. i.e. determining sensitivity capability or to what size contamination is able to be consistently and reliably detected and rejected

Verification would be ensuring this level of control is met. i.e. hourly checks with metal samples to confirm detection and rejection

Monitoring would be evaluating the information the program or process provides. i.e. evaluate customer complaints for metal, review verification logs, evaluate rejected products or failed hourly checks

Most organizations have procedures for verification and monitoring and lack information as to how and why they determined the particular size or type contamination that they are using during their verification checks.

Other processes are validated based on scientific evidence, mainly regarding the chemical and biological concerns.

With physical contamination, there are not any studies, that I am aware of, that covers the broad spectrum of products, system types and sizes, facility and other application variables.

If this is done on an individual basis I am curious what sort of steps the industry takes to satisfy validation requirements



Charles.C

    Grade - FIFSQN

  • IFSQN Moderator
  • 20,542 posts
  • 5673 thanks
1,548
Excellent

  • Earth
    Earth
  • Gender:Male
  • Interests:SF
    TV
    Movies

Posted 10 March 2012 - 12:06 AM

Dear Robert Rogers,

I suggest you try this thread as a starter. Illustrates the background to most of the facets mentioned in yr post. As you will see, it is sort of agreed that there is no single (correct?) solution.

http://www.ifsqn.com...dpost__p__52215

There are some equally detailed items here specifically on validation per se, often for specific standards of interest. If you have one, I suggest a search here in that forum.

Rgds / Charles.C


Kind Regards,

 

Charles.C


mgourley

    Grade - FIFSQN

  • IFSQN Fellow
  • 1,423 posts
  • 1007 thanks
281
Excellent

  • United States
    United States
  • Gender:Male
  • Location:Plant City, FL
  • Interests:Cooking, golf, firearms, food safety and sanitation.

Posted 10 March 2012 - 07:10 AM

What procedure did you utilize to validate your Foreign Body Detection device?


Unless you are building a new line, it's my opinion that you can't "validate" a foreign body detection device within the strict definition of Codex, as the validation has to be done before the equipment is purchased and installed.

What I have done is taken the FDA's "7mm guidance", equipment manufacturers specifications and common industry practice for size of test wands and mashed that into a document explaining why we have that certain metal detector. Is it proper "validation"? Probably not, but it's about the best that can be done until it's time to replace the equipment.

Marshall


Robert Rogers

    Grade - AIFSQN

  • Corporate Sponsor
  • 36 posts
  • 28 thanks
5
Neutral

  • United States
    United States
  • Gender:Male

Posted 16 March 2012 - 05:48 PM

Thanks to all for their input.

mgourley
if a facility was not working toward any certifications in the past and had MD systems in place but now has contract with a customer whom requires certification or they want to comply with changes in the Food Safety Laws would they be obligated to purchase new equipment just to validate the process?



Since the environment and specific location could have an effect on detection capability, it would seem the validation should be performed after installation yet prior to production.



mgourley

    Grade - FIFSQN

  • IFSQN Fellow
  • 1,423 posts
  • 1007 thanks
281
Excellent

  • United States
    United States
  • Gender:Male
  • Location:Plant City, FL
  • Interests:Cooking, golf, firearms, food safety and sanitation.

Posted 16 March 2012 - 06:06 PM

Thanks to all for their input.

mgourley
if a facility was not working toward any certifications in the past and had MD systems in place but now has contract with a customer whom requires certification or they want to comply with changes in the Food Safety Laws would they be obligated to purchase new equipment just to validate the process?



Since the environment and specific location could have an effect on detection capability, it would seem the validation should be performed after installation yet prior to production.



No, they would not be obligated to purchase new equipment. This is what I was referring to in my post. In my situation, the metal detectors were here long before I was and when they were installed, nobody was concerned with "validating" whether or not they were suitable for the purpose.
So now, I can't actually validate the selection of equipment, but I can show that what we have is industry standard and it operates correctly and to the industry standard for the type of metal detection equipment that it is.

Codex is pretty specific in the definition of "validation" and "verification". SQF, not so much. SQF has sort of melded the two definitions as far as I can tell.

My contention was that the only way to "validate" equipment, per Codex, is if it is new equipment, not currently in use.

Marshall


Robert Rogers

    Grade - AIFSQN

  • Corporate Sponsor
  • 36 posts
  • 28 thanks
5
Neutral

  • United States
    United States
  • Gender:Male

Posted 16 March 2012 - 06:12 PM

No, they would not be obligated to purchase new equipment. This is what I was referring to in my post. In my situation, the metal detectors were here long before I was and when they were installed, nobody was concerned with "validating" whether or not they were suitable for the purpose.
So now, I can't actually validate the selection of equipment, but I can show that what we have is industry standard and it operates correctly and to the industry standard for the type of metal detection equipment that it is.

Codex is pretty specific in the definition of "validation" and "verification". SQF, not so much. SQF has sort of melded the two definitions as far as I can tell.

My contention was that the only way to "validate" equipment, per Codex, is if it is new equipment, not currently in use.

Marshall



Understood


Charles.C

    Grade - FIFSQN

  • IFSQN Moderator
  • 20,542 posts
  • 5673 thanks
1,548
Excellent

  • Earth
    Earth
  • Gender:Male
  • Interests:SF
    TV
    Movies

Posted 17 March 2012 - 05:30 AM

Dear Robert Rogers,

Here is a quote from the Codex validation document referred earlier –

IV. CONCEPT AND NATURE OF VALIDATION
Validation focuses on the collection and evaluation of scientific, technical and observational information to determine whether control measures are capable of achieving their specified purpose in terms of hazard control. Validation involves measuring performance against a desired food safety outcome or target, in respect of a required level of hazard control.

Validation is performed at the time a control measure or a food safety control system is designed, or when changes indicate the need for re-validation (see section VII). Validation of control measures is, whenever possible, performed before their full implementation.

(underline is mine)

I guess it is your interpretation when the instrument is formally implemented as a control measure, although logically the validation should surely be done around the time of first usage from a FS point of view. This mini-zigzag is a common pragmatic alternative IMEX.

Here is a previous forum attachment of a “validation SOP” for a labelling control measure vs allergens which quite well illustrates the use of a Validation “trial period” (Canadian origin from memory).

Attached File  validation study example.pdf   152.46KB   133 downloads

Slightly OT but if you are interested, this post and surrounding thread discusses some validation interpretative issues, particularly with respect to SQF standard –

http://www.ifsqn.com...dpost__p__43020

Rgds / Charles.C

Kind Regards,

 

Charles.C




Share this

1 user(s) are reading this topic

0 members, 1 guests, 0 anonymous users